Timestamp Memory usage?


EthanGutmann Oct 27, 2010 06:01 PM

Can anyone tell me how much memory a standard CR1000 timestamp uses? I know it is output as a ~19 byte string, but is it stored internally that way, or is it stored internally as a long (e.g. seconds since 1990).

thanks,
Ethan


jtrauntvein Oct 27, 2010 08:48 PM

The standard time stamp for the CR1000 is 8 bytes long (four bytes for seconds since 1990 and four bytes for nanoseconds into the second.


ChipsNSalsa Oct 27, 2010 08:59 PM

Check out section 9.6.1.2 "DataInterval () Instruction" of the CR1000 manual (http://www.campbellsci.com/documents/manuals/cr1000.pdf) for a fairly detailed description of how time stamps are stored in the datalogger.


EthanGutmann Oct 27, 2010 10:11 PM

Thanks,
I had tried reading through the DataInterval section earlier, and was still confused. That section implies that the datalogger is using 1k/frame, but I don't quite understand what a frame is (I'm writing to the table every 3.5seconds, and it doesn't fill up in ~4 hours as it would if it were writing 1k each time). The relationship between these 1k frames and a record in the datalogger doesn't seem to be spelled out anywhere in the manual.

I'm writing 2 ints and an IEEE float each 3.5 second interval and trying to figure out how long it will last. If it is only storing 8 bytes for the time stamp (and another 4? 8? for the record number) than it should last about a week.

So if I don't use the datainterval() command, but just use
realtime(time)
sample(1,time,NSEC)

'time dimensioned to (1)

Can I save both the record number and the 4 bytes for the milliseconds?

I have a 2GB module on order, but dang, 4MB is frustrating.


Dana Oct 28, 2010 03:18 AM

A frame is just a "chunk" of data in a data table. So, for about every 1K of data that is stored, a timestamp will be stored with it (if data Interval is not used).

One of the simpler ways to get an idea of how long it will take for a data table to fill up is to open LoggerNet's Connect window, click the Station Status button, and view the Table Fill Times tab.

Dana W.

Log in or register to post/reply in the forum.